I have a BMI088 and am trying to get an accelerometer reading.

The data sheet and specs for Bosch Sensortec are here:

The sensitivity per range is this (its a 16 bit signed integer):

  • ± 3 g: 10920 LSB/g
  • ± 6 g: 5460 LSB/g
  • ± 12 g: 2730 LSB/g
  • ± 24 g: 1365 LSB/g

Yet, the formula they give to calculate the force in mg is this:

mg = accel_int16 / 32768 * 1000 * 2^((range register value) +1)

Where the range register has the following values:

  • ± 3g = 0x0
  • ± 6g = 0x1
  • ± 12g = 0x2
  • ± 24g = 0x3

If we do the math on that and plug in the highest value (+32768) and highest range register value (0x3), we get a force of 16000mg or 16g, so a range of -16g to +16, not -24g to +24g.

So what is the correct conversion?

If we do simply, accel_int16 LSB / 1365 LSB/g we get a maximum value of 24g for 32768 which is intuitively what one would expect. Going by the same logic of a linear response, we might also say, each value (LSB) represents a LSB/mg, which would be

32768*x = 24000mg

x = 0.7324

In this case we can take a reading, multiply by 0.7324 to get our mg.

What am I doing wrong or not understanding? Why does the data sheet formula not give me the correct range?

The second part to that question, is, I want to have a scale factor to convert from LSB to to meters/second/second, so intuitively one would think it would be:

scale = 0.7324 * 0.001 * ~9.8m/s/s

Again, if we follow the datasheet formula we get

scale = 0.4883 * 0.001 * ~9.8m/s/s

Am I misunderstanding something fundamental here? Or does a 24g accelerometer only measure in a 16g range for some reason?

706Astor
  • 1,064
  • 1
  • 8
  • 19

1 Answers1

1

I couldn't add a comment (new to the forum) so posting in the answer section. Bosch has since acknowledged this to be an error and updated the datasheet's errata.

They have added a 3/2 factor to the original formula. I am wondering, do we really need to do this way? The LSB$/$g sensitivity just needs to be divided by the register value and with that, if we see, all the subsequent sensitivities are half of the previous value.. so wouldn't the following formula be correct?

${mg = (accel_{int16} * 2^{range\ reg}) / (10920 * 1000)}$

their " * 1.5 " fix, it seems, is reducing to the above formula. What's up with the number 32768?

enter image description here

enter image description here